This initiates the self-calibration routine.
7.
Wait for the self-calibration routine to complete. When the elevation and the
walking belt have completely stopped (approx. 2 minutes), the self-calibration is
finished.
Self-Calibration Routine
a.
b.
c.
d.
Check T2100 Treadmill Level
NOTE:
1.
Use the controlling equipment to verify the T2100 Treadmill elevation is 0.0%.
2.
Check the T2100 Treadmill level with a carpenter's level.
3.
If the T2100 Treadmill is uneven, adjust the feet at the rear of the T2100
Treadmill until it is level.
4.
If you cannot level the T2100 Treadmill safely, move it to another location.
Performing Leakage Tests
The leakage tests are safety tests to ensure that the equipment poses no electrical
health hazards. Use the table below to determine which tests apply to the unit under
test and the maximum allowable leakage currents. For international leakage limits,
refer to the internal standards agencies of that particular country.
If the unit under test fails the leakage tests, do not allow the customer to use the
equipment. Call Tech Support for assistance.
GE Healthcare recommends that you perform these tests:
• Before applying power for the first time
• Whenever internal assemblies are serviced
NOTE:
The accuracy of the leakage tests depends on a properly-wired wall outlet. Do not
proceed until you verify the integrity of the power source.
WARNING:
Total system leakage current must not exceed 300 microamperes.
2021403-120F
The T2100 Treadmill checks the functionality of the elevation limit switches
and the potentiometer by rising to the maximum height limit switch.
The T2100 Treadmill sets the elevation parameters by descending to the
minimum height limit switch and then rising again to the maximum height
limit switch.
The T2100 Treadmill descends to 10% grade and revolves the walking belt
at 3.2 km/h (2 mph) for 30 seconds.
The T2100 Treadmill stops the walking belt and the grade stops at 10%.
Before you check the T2100 Treadmill level, make sure that it has been
moved to its final destination.
T2100 Treadmill
Installation
47