Connect Controlling Devices - GE T2100 Service Manual

Hide thumbs Also See for T2100:
Table of Contents

Advertisement

Installation
metal surfaces are anodized or painted, scrape off a small area in an inconspicuous
area on the aluminum casting, for the probe to make direct contact with the metal.
1.
Connect the AC power cord to the T2100 Treadmill.
2.
Use a digital multimeter to check all the metal surfaces of the equipment as
illustrated below. Make adjustments for any resistance in the test leads.
If the measurements are significantly out of range, check for breaks in the power
cord or in the internal connections within the unit.

Connect Controlling Devices

1.
Attach the appropriate controller cable from the stress system. Use the supplied
interface cable to connect the T2100 Treadmill to the host (for example the
CASE).
NOTE:
NOTE:
2.
Verify that the emergency stop switch is connected.
3.
Connect the AC power cord.
4.
Turn the T2100 Treadmill ON/OFF switch to the On position.
5.
Turn the controlling device ON/OFF switch to the On position.
44
The CASE and MAC 5000ST is connected to the RS 232 Serial Port.
On the CASE system, connect the T2100 Treadmill interface cable in one of
the following ports: COM 1, COM 2, COM C or COM D. The CASE RSS modem
requires the use of either COM 1 or COM 2. Move the T2100 Treadmill
interface cable to COM C or COM D only as necessary.
T2100 Treadmill
2021403-120F