TDU Event /
Error Code
0x16
0x17
2021403-120F
Number
of Red
LED
Flashes
Definition
7
Over-temperature
trip
8
Drive overload (It)
trip
T2100 Treadmill
Check These Items
The ambient temperature of the drive
controller has exceeded the trip level, or
the Motor overtemperature trip input
has been activated.
1.
Allow the unit to cool.
2.
Increase ventilation and reduce
ambient temperature.
3.
Powercycle the T2100 Treadmill
and monitor the T2100 Treadmill
for reoccurrence of this error. (This
may be an isolated incident.)
4.
Determine if any TDU
Troubleshooting Table errors
are also being reported, and if
so, follow the troubleshooting
procedure for the reported errors.
5.
Replace the drive controller.
The drive overload current protection
algorithms have exceeded their limit
and disabled the drive. (The motor/drive
can run with demand currents greater
than their rated value for a period
of time; after that time the drive will
either trip or automatically foldback the
demand current.)
1.
Powercycle the T2100 Treadmill
and monitor the T2100 Treadmill
for reoccurrence of this error. (This
may be an isolated incident.)
2.
Confirm that the motor is not
overloaded or stalling.
3.
Verify connections to braking
resistor are not damaged.
4.
Determine if any TDU
Troubleshooting Table errors
are also being reported, and if
so, follow the troubleshooting
procedure for the reported errors.
5.
Replace the drive controller.
Troubleshooting
97