FRU Replacement
Potentiometer Removal Guidelines
NOTE:
If it is necessary to remove the potentiometer with the intention of reinstallation,
please follow these guidelines to reduce the risk of damage to the potentiometer.
1.
Remove the two mounting plate screws.
Arrows are pointing to the location of the mounting plate screws.
CAUTION:
126
DO NOT apply any pulling force to the potentiometer shaft. If this does
occur, the entire potentiometer harness will require replacement. The
potentiometer is susceptible to damage if axial pulling forces are applied
to the shaft. Do not attempt to remove the potentiometer by grabbing the
potentiometer and pulling it off.
T2100 Treadmill
2021403-120F