Power Cord and Plug

Verify the power cord being used with the treadmill is good:
• Verify that the line, neutral, and ground conductors are properly connected to the
power cord plug and are not short-circuited. Rewire and tighten these, or replace
the power cord, as necessary.
• Failure of the power cord strain relief is very common. Often times users of the
equipment pull on the power cord itself, rather than the power cord plug, to
unplug the unit from a wall receptacle. If in doubt, test for continuity through each
conductor of the power cord connector and plug.
• Perform the
your Country/Local governing safety organization. For international power cords,
refer to the internal standards agencies of that particular country.
• Inspect the power cord, perform safety tests upon receipt of the equipment, per
the recommended maintenance frequency
page
For reference, the following is a list of recommended international wall-side plug
standards for use with unterminated international power cords PN 408930-002
and 408930-004.
Country
Italy
Switzerland
Israel
India/South Africa

Ground Continuity Test

This test verifies that there is continuity (less than 100 mΩ resistance) between all the
exposed metal surfaces, which have the potential to become energized, and the
ground prong on the mains AC power cord. Look for an exposed metal screw, or, if the
2021403-120F
"Ground Continuity Test" on page 43
53.), and each time the unit is serviced.
Standard
CEI 236–16
SEV 1011
SI 32
SANS 164–1
T2100 Treadmill
or test method that is required by
(See "Recommended Maintenance" on
Amperage
16 A
16 A
16 A
16 A
Installation
43