GE T2100 Service Manual page 96

Hide thumbs Also See for T2100:
Table of Contents

Advertisement

Troubleshooting
TDU Event /
Error Code
0x14
0x15
96
Number
of Red
LED
Flashes
Definition
5
Feedback trip
6
Motor overload (I2t)
trip
T2100 Treadmill
Check These Items
Indicates loss of encoder/resolver
feedback and may indicate that the
feedback cable has become detached,
one of the signals has broken, or noise
is present.
1.
Powercycle the T2100 Treadmill
and monitor the T2100 Treadmill
for reoccurrence of this error. (This
may be an isolated incident.)
2.
Check the wiring in the Feedback
cable.
3.
Check the cable screens.
4.
Check for damage to the encoder,
under the motor cap.
5.
Determine if any TDU
Troubleshooting Table errors
are also being reported, and if
so, follow the troubleshooting
procedure for the reported errors.
6.
Replace the drive controller.
The motor overload current protection
algorithms have exceeded their limit
and disabled the drive. (The motor/drive
can run with demand currents greater
than their rated value for a period
of time; after that time the drive will
either trip or automatically foldback the
demand current.)
1.
Powercycle the T2100 Treadmill
and monitor the T2100 Treadmill
for reoccurrence of this error. (This
may be an isolated incident.)
2.
Confirm that the motor is not
overloaded or stalling.
3.
Determine if any TDU
Troubleshooting Table errors
are also being reported, and if
so, follow the troubleshooting
procedure for the reported errors.
4.
Replace the drive controller.
2021403-120F