Troubleshooting
TDU Event /
Error Code
0x0c
0x0d
0x0F
0x80
0x81
0x82
0x83
0xEE
92
Status
Definition and
Detail Description
Temperature
Warning
Warning (*C)<XX>
(where XX equals
the temperature in
Celsius)
Advisory
Self Calibration
Advisory
Belt Stop
Power On
Power Off
Advisory
Time Set
Advisory
Event Log Read
Advisory
Event Log Erased
T2100 Treadmill
Check These Items
The ambient temperature inside
the e-box has reached 55° C
(131° F). The T2100 Treadmill
still operates correctly, but
this warning indicates that
something is causing the
temperature to rise to near-fatal
levels. (65° C is fatal.)
1.
Remove power from the
T2100 Treadmill and allow
to cool off.
2.
Power up the T2100
Treadmill and monitor
the T2100 Treadmill for
reoccurrence of this error.
(This may be an isolated
incident.)
3.
Check the power board
and processor board test
points and verify they are
operating within the test
range. See
"Power Supply
Voltage Checks and Fuses"
on page
100.
4.
Replace the power board
and/or processor board
if they are not operating
within the test range.
Not an error. Indicates self
calibration was initiated.
Not an error. Indicates the belt
was stopped at this time.
The treadmill was powered on.
The treadmill was powered off.
The date and time were set.
The event log was successfully
read.
The event log was erased.