TDU Event /
Error Code
0x04
2021403-120F
Status
Definition and
Detail Description
Over Temperature
Fatal
Error (*C) <XX>
(Halts Current
(where XX equals
Operation)
the temperature in
Celsius)
T2100 Treadmill
Troubleshooting
Check These Items
The ambient temperature inside
the e-box has reached 65° C
(149° F).
1.
Remove power from the
T2100 Treadmill and allow
to cool off.
2.
Power up the T2100
Treadmill and monitor
the T2100 Treadmill for
reoccurrence of this error.
(This may be an isolated
incident.)
3.
Check the power board
and processor board test
points and verify they are
operating within the test
range. See
"Power Supply
Voltage Checks and Fuses"
on page
100.
4.
Verify appropriate
operating environment,
10°C to 40°C (50°F to 104°F)
5.
Replace the processor
board. See
"PCBs
and Power Supply
Replacement" on page
128.
87